I have problems in analysing clindamycin phosphate in a cream.
I have tried some different method, but i always get same result. The analysis result always higher than it should be
( About 10% higher). I have injected the matriks (placebo) and it is not giving any absorbance. Do you have any idea where the 10% came from?

Thanks for your help.

Since you have placebo available, the obvious "fix" would be to always make up your standards in placebo. That way, any matrix effects would cancel out.

You should probably do that experiment (in effect, a "standard additions" run) in any case, and see how your slope and intercept compare to those when running pure standards. My guess is that the slope will be different, which might be explained by slight differences in peak shape due to the matrix changing the integration.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Majestic-12 [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry