Hello,

I'm trying to calibrate PEG on the GPC, my solvent is (deionized) water, but the problem is that i get a very weird chromatograph, with only one very small peak which sometimes is not clear enough to measure.

Do you have any ideas on what the problem might be?

Thanks in advance

What is your detection technique?

Best Regards
Learn Innovate and Share

Dancho Dikov

I called the manufactor and I realized that my column's detection ability was not sufficient for the PEG4000 (MW~3400) that I wanted to measure.
So I guess i have to ask for a new column or use another technique.
Thanks for the quick reply, though..

I meant; what kind of detector are you using?

Regarding the column, it doesn't really matter whether it excludes or includes the analyte. It should be possible to detect a peak somewhere provided a relevant detection technique is applied.

Best Regards
Learn Innovate and Share

Dancho Dikov
4 posts Page 1 of 1

Who is online

In total there are 21 users online :: 4 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot], Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry