We've recently been evaluating the Tosoh SuperSW3000 column (4.6x30) to replace the Superdex 200 10/300 GL columns that we've been using to analyze antibodies and antibody conjugates. The primary reason for the change was the speed, 60 min per injection for the superdex vs 15 min for the SuperSW. The resolution is noticeably better as well. I have noticed the tailing that others have mentioned experiencing with this column, but only with some antibodies. I noticed that Tosoh literature recommends using increased salt concentrations as well as <20% organic solvents to minimize undesired interactions with the stationary phase. To that end I decided to try a different buffer to see if I could improve my results. Previously, we've been using PBS w/ 0.1% sodium azide, pH 6.8. I made the same phosphate buffer, but increased the NaCl conc. from 0.15M to 0.4M, and added 10% ethanol (before pH adjustment). I injected the same antibody (a monoclonal mouse IgG) both before switching buffers, and three times after (I got impatient before the column was fully equilibrated). I'm attaching the relevant chromatograms, but in short what I saw was an increase in retention time, decrease in plate count, increase in tailing, and decrease in area, for the IgG peak. All peaks had better resolution. What is going on here? Are the increased salt and the ethanol having opposite effects? I like the increase in resolution, but the tailing on the IgG peak is getting to be ridiculous. I'd appreciate any theories as to what I've done, and perhaps some suggestions as to how to keep this resolution increase without the increased tailing.
The chromatograms are as follows:
1. Old buffer, PBS w/ azide, 150mM NaCl, pH 6.8
2. New buffer (after about 2 column volumes), PBS w/ azide, 0.4M NaCl, 10% EtOH, pH 6.8
3. Same buffer, after 4 column volumes
5. Same buffer, after many column volumes
Flow rate: 0.35 ml/min
Sample Conc: 2 mg/ml
Injection vol: 5 ul
Detector: UV @ 280 nm
Any ideas?
Scott Horn