Sugar Separation by UPLC/ELS
Posted: Fri May 28, 2010 3:53 pm
Hello science comunity.
Im developing a method for separate and quantify Fructose, Glucose, Sacarose,, Maltose and Lactose in foods.
Im using a UPLC with an ELS, but when I run the sugars standars individually I see two peaks for each sugar, and Im not suppose to see more than one peak for each sugar standars. Im not seeing this when I run it on the HPLC. My task is to migrate from HPLC to UPLC.
Im using the Amide BEH Amide 1.7um, 1.0x100mm Column. the mobile phase I using is 90%ACN-10%H2O with a gradient to 75%ACN-25%H2O in 8 minutes. the flos is 0.3mL/min and tehe column temperature is 35C. The detector setting are: Drift tube temperature 60C, Nebulizer at 12C (cooling), Gas Pressure N2 at 50psi. Gain 200.
The sample and standars used was 60%ACN/40%H2O. The volumn inyection is 1.3uL and the concentration of the standar being tested is 3mg/mL.
As the sugars can convert from cis to trans, but the Sacarose looks good. The simple sugars are the ones thant shows two peacks. Probably contamination fo the standar....well Im using fresh standars from Merck up to 99% purity.
Thanks for your opinions.