Sugar Separation by UPLC/ELS

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
Hello science comunity.

Im developing a method for separate and quantify Fructose, Glucose, Sacarose,, Maltose and Lactose in foods.

Im using a UPLC with an ELS, but when I run the sugars standars individually I see two peaks for each sugar, and Im not suppose to see more than one peak for each sugar standars. Im not seeing this when I run it on the HPLC. My task is to migrate from HPLC to UPLC.

Im using the Amide BEH Amide 1.7um, 1.0x100mm Column. the mobile phase I using is 90%ACN-10%H2O with a gradient to 75%ACN-25%H2O in 8 minutes. the flos is 0.3mL/min and tehe column temperature is 35C. The detector setting are: Drift tube temperature 60C, Nebulizer at 12C (cooling), Gas Pressure N2 at 50psi. Gain 200.

The sample and standars used was 60%ACN/40%H2O. The volumn inyection is 1.3uL and the concentration of the standar being tested is 3mg/mL.

As the sugars can convert from cis to trans, but the Sacarose looks good. The simple sugars are the ones thant shows two peacks. Probably contamination fo the standar....well Im using fresh standars from Merck up to 99% purity.

Thanks for your opinions.
jcastaneda

you might have a mismatch between diluent and your mobile phase, you might be overloading the column ( 1x100 mm is kind of small) or you formed a void in the column
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com

This separation can be easily performed on Unison UK-Amino:
http://www.imtaktusa.com/site_media/fil ... TI325E.pdf

Follow the above method using Unison UK-Amino and all your problems go away...

What you observe is absolutely normal, and the solution is equally well known. I will explain below, but you can look at the Waters applications as well.

Sugars form anomers, which are two forms of the same molecule that can interconvert when the ring opens. This is the reason that you get two peaks for your sugars. Saccharose does not do that, this is why you get a single peak.

In order to get a single peak out of the two peaks, it is necessary to speed up the equilibrium between both forms. This can be done by adjusting the pH to alkaline and elevated temperature. For your application, we used 0.2% triethylamine (TEA) at 35 degrees, or 0.05% TEA at 90 degrees.


In order to get a single peak out of the two peaks, it is necessary to speed up the equilibrium between both forms. This can be done by adjusting the pH to alkaline and elevated temperature. For your application, we used 0.2% triethylamine (TEA) at 35 degrees, or 0.05% TEA at 90 degrees.
...and column lifetime under such harsh conditions???

Of course, if this were a silica column, if would go belly-up in a few hours. However, the BEH Amide column is based on the ethyl-bridge hybrid material, which is good to pH 12.

I would not give a presentation at the HPLC conference on stuff that has not been proven to work, and the material described in my post cam straight from my lecture.

I can send copies of the lecture to anybody interested, including you, Bryan. You might learn something...

Hi Uwe -

Thank you for the kind offer - but I think I've got a pretty good understanding of UPLC and BEH technology.
I do my homework :lol:
7 posts Page 1 of 1

Who is online

In total there are 19 users online :: 3 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry