LC pump performance

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
An HP 1050 LC pump is behaving strangely. When it pumps a 0.2% isopropanol in cyclohexane mix there is a +/- 50+ psi pressure swing. When pumping tetrahydrofuran the swing is +/- 3 psi. Pump has been thoroughly rebuilt (seals, check valve, purged, etc) and seems to operate normally. The IPA/CH mix is run through a de-gasser.

Is there something about the presence of dissolved gases or compressability between the two liquids systems that could explain the pressure instability?

Dear erlex,

The last time I used an HP1050 I was delighted to get 3psi of ripple. Your pump is fine!

Cyclohexane, on the other hand, is extremely compressible. I don't know what factor to use, but try increasing the compressibility compensation on your pump.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 4 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], Google [Bot], Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry