EI or CI in GC/MS?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hi All,
My company is buying a GC/MS for residual solvent and extractable/leachable analyses. Does anyone have an opinion on the ionization if both EI and CI are needed, or EI should suffice in most cases?

Thanks in advance. :wink:
NomadBuck

From my experience working with EI and CI, Negative CI is really useful if you want to look at very low (ppb or lower) levels of halogenated and certain other highly electronegative compounds. If these types of compounds are not of interest to you or you don't need really low LODs for them then EI mode alone will probably be suitable for your needs.


Also, in the case of Agilent's 5975 (possibly other vendors as well) you can always buy the 5975 EI and then pay them later for the CI upgrade.

Residual solvents and leachables are things that need to be legally defensible in most cases, and there are no standard methods for these assays using CI. Unless you are willing to do an extensive validation, I would not recommend being a pioneer.

Again, Thank you for your insight, jh1 and Ron.
NomadBuck
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry