EFC type-4 Varian GC-3800 malfunction

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Dears, EFC typ e 4 is installed with hydrogen as carrier, setpoint of pressure is 18 psi and flow rate is 60ml/min. I am getting zero pressure and flow of 79ml/min. What could be the possibilities. GC is not ready and I cannot run any sample.

The first thing that springs to mind is a massive leak. Check with a leak seeker. Is this a capillary split-spltless setup ? or something different ?

Peter
Peter Apps

EFC is installed on packed column, a simple application.
Well I will try to detect any leakage.....Thnx

Is this the EFC module with only mass flow control ? If so it makes troubleshooting easier. Presumably with a packed column you have no split and no septum purge, so the gas flow goes only though the column.

With hydrogen at 60 ml min as carrier I presume that your detector is TCD, so simpler yet, with no fuel or make-up gasses.

Measure the flow at the detector using a flow meter calibrated for hydrogen - a bubble flow meter is simplest.

What flow do you get ?

Peter
Peter Apps
Dear, It was faulty EFC causing the problem. Problem seems to be with the pressure regulation/sensing. It was giving a flow of 100 ml/min after a MS column without showing any column head pressure. No leakages were found. Thanks for your support.
5 posts Page 1 of 1

Who is online

In total there are 97 users online :: 0 registered, 0 hidden and 97 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 97 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry