Air leak in the system.
Posted: Thu May 20, 2010 3:47 pm
by BCBC
I isolated the MSD (5973 with turbo) from the GC to find an air leak.
An air/water check gives me the following.
18/69 = 55.50 % Water
28/69 = 6.75% Nitrogen
32/69 = 2.49 % Oxygen
44/69 = 0.88 % Carbon Dioxide
28/18 = 12.85 % Nitrogen/Water
Any thoughts? I had about 13 % water before I isolated the MSD.
So I know I introduced air with water when I capped the interface but this was about 3.5 hours ago and I'd expect the water % to go down...maybe not it is humid here.
Re: Air leak in the system.
Posted: Thu May 20, 2010 6:01 pm
by astroguille
Hi!
Check the seals (vent valve, lateral seal). If OK, check the o-ring of the interface.
Good luck!
Posted: Thu May 20, 2010 7:48 pm
by BCBC
Thanks for the information.
I replaced the side door o-ring. have replaced the interface o-ring.
Never had a leak indication in those two places but replaced them anyway.
Posted: Fri May 21, 2010 12:07 am
by Don_Hilton
If the nitrogen is OK (or close to OK), let the instrument pump. Water adsorbs to surfaces and comes free slowly. If you have capped off the MSD, you have no carrier flow to sweep water out of the source. Expect it to clear more slowly.
Posted: Sat May 22, 2010 12:42 pm
by BCBC
I'll know more on Monday after a weekend of pumping down.
Posted: Sat May 22, 2010 2:38 pm
by haiedc
You should connect the column to the MSD for the carrier gas to sweep out water trace. Anyway, leak (if any) from the interface, inlet (column) can add air and water into the msd.
Posted: Sat May 22, 2010 3:09 pm
by Don_Hilton
There are times when it is appropriate to plug off the transfer line to look for a leak. It is a pain because the gas flow conditions have changed. But, it can help to confirm that a leak is in the mass spectrometer.
It is rare when there is a leak in a mass spectomter. Typically is a seal that you have had apart. On ocassion, it is something like a pass-through that has cracked. (A perfect end to a bad day!)
My personal preference for tracking down a leak in the mass spectrometer is to reopen the seals that I have had apart, clean them, and reassemble them (including the seal for the vent!) -- with the exception of copper o-rings. With copper o-rings, if you always use new o-rings and properly torque the bolts, it is next to impossible to fail to seal the joint. If the leak goes away, problem solved.
Posted: Wed May 26, 2010 4:20 pm
by BCBC
Monday I had 8 % water, 5% nitrogen and 2 % oxygen.
It looks better. I replaced the ion source as 219 was getting pretty low in abundance an let it pump down for a few hours.
Next check will be on Thursday. I'm off today.
Thanks for the help.
Posted: Mon Jun 07, 2010 2:58 pm
by BCBC
Changing out the PFTBA helped a lot with the water level.
Reconnecting the column with helium flow also swept out water.
This morning I have nitrogen at 2.5 %, Oxygen at 1.36 % and
water at 2.4 %.