Question on calibration & solvent strength .

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi Folks ,

I am using isocratic RP-HPLC , 65% methanol /water , flow rate 1ml/min ,column 250mm 4.6 id . i build up my calibration graph and use it for more than one year which was very time consuming . My question is , if i increase the solvent strength to 75% with keeping every things the same will i be able to use the same calibration curve with relating any unknown conc. from 75% to the calibration of 65% .In other wrod , i change only the mobile phase strength .

The shortest answer, that is still adequate is: No!
You'll have to document the validity of your results once again.

Another story is, how you did document your current procedure (i.e. calibrating once a year)?

Best Regards
Learn Innovate and Share

Dancho Dikov
2 posts Page 1 of 1

Who is online

In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry