metal contamination causing imp peaks at high pH eluent?
Posted: Fri May 14, 2010 10:01 am
My problem started when I installed a new column….
Using an Xbridge 150 x 46 mm, 3.5 um column at pH 9.5 NH4Cl-acetonitrile mobile phase, I observed two (extra) very repeatable peaks in both standard and sample solutions. These peaks however, were not observed in any blank solution. Have you ever seen anything like this? Could the analyte be complexing with metals in the stationary phase? Also, the standards/samples themselves had been exposed to Pt/alumina catalysts, and in addition the main peak exhibited increased tailing.
I did not see the two peaks on the previous column and peak tailing was also less -but I had run with 0.1%H3PO4-acetonitrile mobile phase first, before I switched to the high pH method.
Keeping that thought I eluted (95:5) 0.1%H3PO4:acetonitrile through the new column overnight and the next day when I switched back to the high pH mobile phase, the two peaks were gone.
I would be very interested in any insight/feedback you might have.
Thanks!
Thanks!