Hello,

I am using a 0.0015 M H2SO4 mobile phase with a Rzex-ROH 7.8*300 mm column coupled to a conductivity detector, seperation of fluoride is excellent from water dip. Will quatitative determination fluoride work with this method?
I was just thinking that HF is not a strong acid and the dissosiation of fluoride ion will vary with concentration and pH, thus not giving a linear calibration curve. What will approximatly be the linera range for this set-up. Must I switch to an alkaline mobile phase.

Thanks for your answer.

With that dilute a mobile phase, the potential for problems definitely is there. The only way to tell for sure is to run a calibration curve and see how linear it actually is.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

I think that you can change the moble phase to 0.0015 M Na2SO4, and the problem maybe sovled.

But the sulfate will be difficultly eluted if you use ion-exchange column. So my suggestion is that you'd better use Sodium hydroxide or potassium hydroxide.
If your column is ion-exclusion column, both separation and linearity will be better with 0.0015 M sulfuric acid.
The two modes have different separation mechanism.

Dear Daggio

I do not know that column. But eluent and column diameter make me assume that you are running an ion exclusion application.
Ion exclusion is mainly used for weak acid like organic acids. As HF is a relatively weak acid it is retained and quantification is possible.
I assume a working range from the e.g. 50 µg/L up to the mg/L range.

Changing the eluent pH is not required unless the column is an anion exchange column. In that case an alkaline eluent combined with suppression might increase the sensitivity of the method.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland

Hello,

Thanks for your input.

Yes, its an ion-exclusion column and application.

I was just assuming that if you inject eg. 0.1 mg/mL or 0.2 mg/ml fluoride you will end up with pretty much the same responce on the conductivity detector as the conductivity is dependent on concentration of F-.
The detection limit is not an concern. I am actually running a limit test and I dont want to see more than approx. 0,45 mg/ml fluoride in my product. I was wondering if this concentration is actually too high using my described method.
Was that clear? :-)

Thanks.

Yes, I assume that you need to dilute your sample e.g. 1 : 100.
Just give it a try.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland

Side comment: For quantitation of just F-, you could use a flouride ion selective electrode
8 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry