krickos » Fri May 07, 2010 3:06 pm
Probably it's a case of overloading.
Agreed, most of the settings are pretty much standard, the volatiles inlet I use have a volyme of like 32-37µl compared to a 0,250ml still works fine.
0,15mg/ml water can cause 2 things:
A. Overload as stated fact is that 5 time 0,15mg gives 7500µg intotal in vial (which is pretty high), due to 1. using water as main solvent, " headspace oven temp way above methanol you get most methanol in gasphase. 2. Reducing column dimesion reduces your sample load capacity.
B. Rodney mention the water vapour issue, though myself I have only encountered it above 80°C in HS oven. (then again most method I have developed have been based on DMF as solvent).
So as suggestion EP/USP have a set of different typical HS parameters, as resulution is an issue with the 0,53mm id column, pick the setting with 60°C for headspace oven and see if that works better (less solvents in vapour) or validate your own method.
You do not seem to have much lean way otherwise as flow 2ml/min ~30cm/s and column oven temp is 40°C with column that is pretty much optimized.
Ohh one last thing, EP if recalling right mentions that sample/standard preparation might have to be adjusted compared to their default methods.
USP do not have that added at least yet.