Posted: Fri Apr 24, 2026 8:52 pm
by jbinksi
When we power cycled the mass spec chemstation now sees the G6125B as a 5973N system. Can I reload the firmware for the system?

Re: G6125B Recognized as 5973N

Posted: Tue Apr 28, 2026 3:00 am
by linda68
I've encountered a similar case before; the problem wasn't with the firmware, but rather ChemStation loading the wrong old configuration (5973N). Usually, I would delete the old configuration, recreate the instrument from scratch, and then reconnect. Sometimes, this fixes the problem without needing to touch the firmware. Among Us