"phantom" peak
Posted: Sun May 09, 2010 8:27 pm
Hello everybody.
I found plenty of very clever men here so I hope you can help me with problem I have been faced with.
We test purity of a very sensitive substance. Unfortunately I cannot send structure of the product.
Chromatographic conditions are:
Column: Silica based C-18, 250x4.6 mm
Mobile phase: Acetonitrile/borate buffer, pH approx. 10
Phase used for dilution: Acetonitrile/20mM solution of glycine (added for better stability of product)
Substance is very sensitive and unstable in neutral or acidic conditions. As stabilizing agent NaOH is added into product (crystalline powder).
Sample is weighted, diluted in "dilution phase" and injected.
Solvent peak elute at 2 min.
Main product elute at 12 min.
Some batches contain impurity which can be seen at 2.7 min. (my "phantom" peak) and it is source of my headaches.
The "phantom" peak has been found only in several batches. When this batch is re-tested its amount is the same. If the same sample is injected after 15 minutes the peak disappear.
Till now you can say OK, you have unstable contaminant, but...
- UV spectrum is the same as main compound (99% match - Chemstation)
- It seems that product of decomposition of my "phantom" peak is the same as product of decomposition of main compound
- When I did semipreparative injection, quickly re-analysed collected "phantom" peak I found only two peaks: product of decomposition and ...
... main compound!
My result is that small amount (up to 0,1%) of main compound is teared-off during injection and elute practically unretained. But why is it observed only in some batches and on the same level?
Have you seen such strange chromatographic behaviour? Do you have any other explanation for me? So far I have thought that I know quite a lot about chromatograpy but somebody on high kicked my self-confidence ...
Thank you for comments and forbearance to my english.

I found plenty of very clever men here so I hope you can help me with problem I have been faced with.
We test purity of a very sensitive substance. Unfortunately I cannot send structure of the product.
Chromatographic conditions are:
Column: Silica based C-18, 250x4.6 mm
Mobile phase: Acetonitrile/borate buffer, pH approx. 10
Phase used for dilution: Acetonitrile/20mM solution of glycine (added for better stability of product)
Substance is very sensitive and unstable in neutral or acidic conditions. As stabilizing agent NaOH is added into product (crystalline powder).
Sample is weighted, diluted in "dilution phase" and injected.
Solvent peak elute at 2 min.
Main product elute at 12 min.
Some batches contain impurity which can be seen at 2.7 min. (my "phantom" peak) and it is source of my headaches.
The "phantom" peak has been found only in several batches. When this batch is re-tested its amount is the same. If the same sample is injected after 15 minutes the peak disappear.
Till now you can say OK, you have unstable contaminant, but...
- UV spectrum is the same as main compound (99% match - Chemstation)
- It seems that product of decomposition of my "phantom" peak is the same as product of decomposition of main compound
- When I did semipreparative injection, quickly re-analysed collected "phantom" peak I found only two peaks: product of decomposition and ...
... main compound!
My result is that small amount (up to 0,1%) of main compound is teared-off during injection and elute practically unretained. But why is it observed only in some batches and on the same level?
Have you seen such strange chromatographic behaviour? Do you have any other explanation for me? So far I have thought that I know quite a lot about chromatograpy but somebody on high kicked my self-confidence ...
Thank you for comments and forbearance to my english.