A funny question
Posted: Sat May 08, 2010 6:33 am
by brightsun
Hello everybody,
Now i'm going to analyse Fenobucarb ( a herbicide compound),
But i have a funny question in making mobile phase :
In a document i have " The mobile phase was comprised of 87% of water, 13% of acetonitrile,and 0.1% of acetic acid."
And i make this mobile phase by getting 87ml H2O, 13 ml ACN and 0.1 ml acetic acid. All products belong Merck, HPLC grade.
Is it right solution ?
Thank you for your reply !
Posted: Sat May 08, 2010 7:41 am
by thohry
I should think it's in weight ratio.
Posted: Sun May 09, 2010 9:21 am
by brightsun
Thanks thohry
May someone give me any solution ?
Thanks !
Posted: Sun May 09, 2010 11:52 am
by Don_Hilton
Would it be reasonable to try the solution you made and see if it works? If the separation is not what you want, adjustement to thohry's mixture would be somethign to try? But, the two mixtures should be fairly similar to each other.
Beyond that, the best idea is to try to track down the author of the document or someone who has used the procedure sucessfully.
Posted: Sun May 09, 2010 3:38 pm
by aysal
hi,
you may use 2 mobile phase: mobile phase A: Water containing 0.1 % Acetic acid and Mobile phase B: Acetonitrile containing 0.1 % acetic acid. Then you may use 87:13 composition (final acetic acid is always 0.1 %) or different composition until you get good resolution. Therefore you don't need to prepare new mobile phases for every step. just inject your std and see what you get.
Ayhan
Posted: Sun May 09, 2010 3:39 pm
by aysal
I hope you do not have isocratic pump
Ayhan
Posted: Mon May 10, 2010 8:44 am
by HW Mueller
brightsun´s question is not funny, but the label of % is. Most likely the authors meant mL, not %. They could have also meant to have used g. Any attempt to justify the use of % here is bound to fail.
This is a typical example of sloppy reporting.
Posted: Mon May 10, 2010 6:59 pm
by JGK
My best guess would be a some point the analyst made a water:acn (87:13 v/v) isocratic mobile phase tried it but got crappy results and decided to add a bit of acid to improve things
It probably succeeded and writing water: ACN:HAc (87:13:0.1v/v/v) is virtually equivalent 86.9: 13.0: 0.1 v/v/v (where the components add up to 100)
Posted: Wed May 12, 2010 2:17 am
by brightsun
Thank you for all !
I think they use gradient pump.
Firstly, they mix 87% Methanol and 23% H2O in one bottle (A). The another bottle contains CH3COOH (B). After that, they use gradient pump with component : 99.9% A and 0.1% B
I think that.
If my way is not okay, i will try making all your suggestions.
Thanks !
Posted: Thu May 13, 2010 6:46 am
by Rolandas Plausinaitis
Thank you for all !
I think that.
Thanks !
Certainly not. Precission of mixing 99.9/0.1% in gradient pump will be wery bad not speaking of damage made to the pump by the acid.
It was most likely meant 87:13:0.1v/v/v as suggested by JGK.
Taking into account decrease of volume after mixing 87 ml Methanol and 23 ml of Water addition of acid might even make it again 100 ml. :wink:
On the other hand I think in most cases you will not see substantial difference when using mixtures of 99.9+0.1 or 100+0.1 ml
Posted: Thu May 13, 2010 7:26 am
by brightsun
Thanks Rolandas Plausinaitis !
Posted: Thu May 13, 2010 1:57 pm
by Consumer Products Guy
I agree with the others who think it's by volume. This just shows why these types of instructions absolutely need to be written so they're definitive and not open to interpretation. Besides, automated HPLC mixing valves mix by volume.