I hope that someone has more experience and could help me with my issue, which I’m not able to resolve it recently.
I have a LC-MS/MS method for determination of some organic acids in environmental samples, using deuterated internal standards. For the aromatic dicarboxylic acids I use phthalic-D4 acid as IS and for the aliphatic dicarboxylic acids – succinic-D4 acid.
Some of the chromatographic and MS parameters:
- Agilent 1100 LC coupled with AB 4000 QTRAP
- Xbridge Amide 100x3mm, 3.5um column
- Gradient elution using mobile phase containing ACN/H2O with 10 mM ammonium acetate (adjusted to pH=5 in water, before mixing)
- Column thermostating with means of thermostated water bath (25C)
- Negative Single Reaction Monitoring MS/MS method (or (-)MRM, according to the Analyst software)
During FIA-MS optimization of the ion source (Turbospray – ESI source) I can clearly detect and optimize the parameters for both natural and deuterated acids. The problem occurs when I connect the column and run the LC-MS/MS analysis. Phthalic-D4 acid peak shows in the chromatogram (coeluting with the phthalic acid peak) which is ok, but I cannot detect the succinic-D4 acid, although the succinic acid peak is clearly present and symmetric. The SRM (or MRM) transition that I use for succinic-D4 acid is 121.0->59.0 (from syringe injection – MS optimizations of compound dependant parameters).
The stock standard solution of succinic-D4 acid was prepared once in methanol and once in acetonitril with no change/improvement in the chromatogram. The final dilutions of the standard were made in pure ACN. Changing the mobile phase additive from ammonium acetate to ammonium formate and playing with the column temperature didn’t resolve the problem of the missing succinic-D4 peak.
Before I posted this problem, I have searched and read a similar topic on this Forum, where on column degradation of deuterated standard was suggested as possible cause of the lowered peak intensity in chromatogram (the deuterated standard was different from this situation) The link: viewtopic.php?t=3674&highlight=deuterated
Also on the XIC I’m observing higher background intensity for the succinic-D4 transition. I was thinking to try LC-UV to see if this deuterated standard elutes from the column and is detected as the succinic acid is, but its presence during FIA-MS tells me that it is not a MS issue, but chromatographic. And here I'm stuck
Deuterated standards were purchased from Sigma Aldrich.
Did you have similar problems like this. What is your experience? What could be the problem?
I appreciate your time and help.
Best regards