Vitamin C assay-repeatability issues
Posted: Fri May 07, 2010 5:43 pm
Thanks for looking any input would be appriciated
As stated Im having issues with repeatability of standards for a Vitamin C assay. My sequence consists of three injections of standard followed by three injections of sample and finished with three more standard injections. Im getting %RSD of 7-15% for the standards with the area counts increasing over time. I got individual vials for each injection (eliminated evaporation as cause). Filtered with individual syringe filters for each vial (eliminates filter build up or blow-through). I read on here about Ascorbic acid reacting with the glass, how plausible is this theory?
Any other ideas would be appricated.
Assay parameters
mobile: 10:90 ACN: H2SO4 4mM
column: bio-rad aminex HPX-87H
runtime: 30 min
diluent 0.025% m-phosphoric acid
wavelength:210nm
inj volume: 30 ul
As stated Im having issues with repeatability of standards for a Vitamin C assay. My sequence consists of three injections of standard followed by three injections of sample and finished with three more standard injections. Im getting %RSD of 7-15% for the standards with the area counts increasing over time. I got individual vials for each injection (eliminated evaporation as cause). Filtered with individual syringe filters for each vial (eliminates filter build up or blow-through). I read on here about Ascorbic acid reacting with the glass, how plausible is this theory?
Any other ideas would be appricated.