Hi everybody! I am trying simultaneous quantification of vitamins A, E and D3. I tested them in spectrometry UV to know the best wavelenght for the LC analysis and I found different wavelengh for each one. Now, I have a doubt: have the wavelengh choosed for run LC an influence on the peak areas of the vitamins? How I chose the best wavelengh for them together? Thank you in advance

Below is are fat soluble vitamins separated on Cadenza CD-C18 (UV @ 280nm):

http://www.imtaktusa.com/site_media/fil ... TI121E.pdf

Since the question, "have the wavelengh choosed for run LC an influence on the peak areas of the vitamins?" has not been addressed:
Of course, if you choose a wavelength where your analyte does not absorb you don´t get a peak at all.
If your detector allows this you can change the wavelengths to the max lambda of each compound in the region where it elutes. According to Bryan a good compromise is to use 280 nm for all. In any case you have to calibrate.
3 posts Page 1 of 1

Who is online

In total there are 323 users online :: 1 registered, 0 hidden and 322 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 322 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry