Welcome to the forum
Drift during a batch is common with equilibrium headapce analysis (and with other injection techniques).
Nobody can tell you anything useful without knowing the scale of the problem and the details of the method.
Peter
Thank you for your response, Mr. Apps. Samples that are in the latter portion of the run experience a positive bias nearly 7%. (e.g. a NIST traceable CRM of 300mg/ml quants as 320mg/ml). Any of your insights would be greatly appreciated. Please let me know if there's another parameter that you're interested in, I'll be more than happy to provide.
Internal Standard (I.S.): n-propanol (0.5 v/v%)
Headspace vials: Restek 20ml, septa PTFE/silicone
Samples are prepped at a 1:10 dilution--100ul of sample, 1000ul of I.S.
Calibration model specs: currently on a 6-point calibration curve, working range is 10mg/ml up to 400mg/ml. Quadratic curve, no forcing, no weight.
Instrument A
Headspace
Oven Temp. : 70.0 C
Sample Line Temp. : 150.0 C
Transfer Line Temp. : 160.0 C
Shaking Level : 3
Multi Injection Count : 1
Pressurize Gas Pressure : 14.7 psi (nitrogen)
Equilibrating Time : 10.00 min
Pressurizing Time : 0.50 min
Pressure Equilib. Time : 0.10 min
Load Time : 0.50 min
Load Equilib. Time : 0.10 min
Injection Time : 0.50 min
Needle Flush Time : 0.50 min
GC Cycle Time : 4.60 min
Injector/GC
Injection Mode : Split
Carrier Gas : H2
Flow Control Mode : Linear Velocity
Pressure : 7.0 psi
Total Flow : 52.4 mL/min
Column Flow : 3.15 mL/min
Linear Velocity : 55.0 cm/sec
Purge Flow : 2.0 mL/min
Split Ratio : 15.0
Isothermal program, 35.0 C
RTX-BAC Plus 1 column
FID
Temperature : 300.0 C
Sampling Rate : 200 msec
Stop Time : 4.00 min
Delay Time : 0.00 min
Subtract Detector : None
Makeup Gas : N2
Makeup Flow : 24.0 mL/min
Detector Flow Const Mode : No
H2 Flow : 32.0 mL/min
Air Flow : 200.0 mL/min
Instrument B
Headspace
Oven Temp. : 70.0 C
Sample Line Temp. : 150.0 C
Transfer Line Temp. : 160.0 C
Shaking Level : 3
Multi Injection Count : 1
Pressurize Gas Pressure : 14.7 psi
Equilibrating Time : 10.00 min
Pressurizing Time : 0.50 min
Pressure Equilib. Time : 0.10 min
Load Time : 0.50 min
Load Equilib. Time : 0.10 min
Injection Time : 0.50 min
Needle Flush Time : 3.00 min
GC Cycle Time : 4.60 min
Injector/GC
Injection Mode : Split
Carrier Gas : H2
Flow Control Mode : Linear Velocity
Pressure : 5.1 psi
Total Flow : 36.2 mL/min
Column Flow : 2.26 mL/min
Linear Velocity : 41.2 cm/sec
Purge Flow : 0.0 mL/min
Split Ratio : 15.0
Isothermal program, 35.0 C
RTX-BAC Plus 2 column
FID
Temperature : 300.0 C
Sampling Rate : 200 msec
Stop Time : 4.00 min
Delay Time : 0.00 min
Subtract Detector : None
Makeup Gas : N2
Makeup Flow : 24.0 mL/min
Detector Flow Const Mode : No
H2 Flow : 32.0 mL/min
Air Flow : 200.0 mL/min