The matrix can have interferences , for example N2 and CO are both mass 28, and as Don has pointed out the separation on a standard capillary column may not be too good unless you go subambient. Some mass spectrometers can handle the flow from a 0.32 mm id column, and this would allow the use of a PLOT column IF a particle trap is used. This will give a better gas separtion.
I usually advise against the use of a mass spectrometer for the analysis of light gases, it can be done, but it is a lot of work, and is generally much easier to use a GC with a TCD or a discharge ionization detector.
Is there a possibility of adding a detector and second injection port to the GCMS you have access to? That might be the best solution to your problem