FID response of HCHO and HCOOH
Posted: Mon May 03, 2010 9:44 am
by haiedc
Hi all,
According to ECN table, HCHO and HCOOH have no response in FID. Anyone has ever tested this? No peaks or just tiny peaks?
Posted: Mon May 03, 2010 1:48 pm
by Consumer Products Guy
There will be response. However these are both somewhat reactive/polar and do not give good peak shapes when trying to assay "as is".
Posted: Mon May 03, 2010 2:30 pm
by AICMM
haiedc,
Response but very, very poor. Better to do this with argon ionization detector (with my well known bias) or, I have it from a very reliable source, you can use a methanizer and FID combination. Thick film PDMS will work pretty well for this application.
Best regards.
Posted: Mon May 03, 2010 3:21 pm
by chromatographer1
No response is wrong, or at beat an exaggeration.
But remember even WATER can be seen as a small peak on an FID, so formaldehyde and formic acid giving a small response is no surprise.
(and no one argue with me on this on theoretical grounds as I did residual solvent determinations for decades with the first peak always being the water peak using porapak Q packed columns. Water (steam) doesnt ionize in the flame necessarily but it does change the electron flow between the FID elctrodes and give a 'response'.)
A TCD, an argon or helium ionization detector will detect both analytes with excellent sensitivity. Buy what you need when you need it.
rodney george
consultant
Posted: Tue May 04, 2010 1:54 am
by haiedc
rodney george
consultant
Thanks all for replying. Talking about H2O response, I have never seen, but CS2 really gives response, and very low providing that it is used as a diluter.
Posted: Thu May 06, 2010 4:03 pm
by larkl
Th FID is essentially a carbon counter. With these molecules a much smaller fraction of the MW is carbon, compared to a larger similar molecule. One reason for the poor FID response.
Posted: Tue May 11, 2010 11:08 pm
by bhahn
Rodney is correct, you will see peaks, small ones, but you will also see really bad chromatography, as these are extremely polar compounds.
Good luck,
Bill