- Posts: 1
- Joined: Mon Sep 29, 2025 11:37 pm
Values from this morning show abundance of water at 25,880 (18/69=3.07%) and nitrogen 41,344 (28/69=4.90%) and oxygen 15,124 (32/69=1.79%) after purging the calibration valve. We did another 4min purge and the values are water 24,064 (18/69=4.06%), nitrogen 22,216 (3.75%), oxygen 5,215 (32/69=0.88%). Another 4 min purge shows water 10,035 (18/69=3.27%), nitrogen 5,559 (28/69=1.81%), and oxygen 1,611 (32/69=0.53%). However, when you let the MS sit for 10 minutes and DO NOT purge the calibration vial beforehand, the values from an air and water check now read: water abundance 28,368 (18/69=2.73%), nitrogen 56,152 (5.40%), oxygen 12,624 (32/69=1.22%).
We didn't see much nitrogen or oxygen before venting the MS and messing with the vent valve and calibration bulb but it seems to be venting out. Did a manual tune while spraying Argon and saw no signs of a leak. Given the low abundance of water (and N, O, etc.) during a manual tune with the PFTBA valve CLOSED, we believe the water is coming from the valve or bulb itself. Has anyone else experienced this? Any advice or recommendations? We are planning to replace the PFTBA with fresh stuff (it's probably never been changed... although it is at the correct height), but besides that we are puzzled on what to do next...