ghost peaks, small standard = inaccuratly high analytes
Posted: Fri Apr 30, 2010 8:44 pm
I'm doing quantification of VFAs (acetic, propionic, isobutyric and butyric) in paper in aqueous solutions. The samples are acidified with HCl and internal standards are either crotonic or pivalic acid.
My parameters are:
Focus GC FID
Column; Stabilwax-DA column (Restek 30m-0.25mm ID-0.50μm df)
Four Temp; 100°C to 190°C @ 5°C/min, 240°C @10°C/min
Inj. temp; 200°C split 7:1
Carrier; N2 constant pressure 225kPa
Det. Temp; 200°C
The problems are these, ghost peaks on blank injections and decreased standard peak area for samples as compared to calibration injections.
Problem #1 a blank of water has peaks for all the analytes and the standard. running with machine without a physical injection shows no peaks. The ghost peaks seem to be a problem of residue somewhere, but where? If I replace the septum, liner, syringe, solvent vials and solvent, it goes away. But with just one injection (10-100ppm analyte) they're back. It is not economical for me to replace all this each injection. I did some calculations and my vapor volume is 1/3 that of the liner volume so I'm doubting backflash is causing the residue. I'm rinsing the syringe 5 times with water and 5 times with acetone/methanol before and after injection and three times with sample. The peaks will minimize if I inject water many times, but who has the time for that? Ideally I'd like the analytes and standard to stop sticking to whatever they're sticking to.
Problem #2 I can run several calibration standards and get a pretty decent calibration using the internal standard method. The peak area is pretty stable. Then I inject a sample and the standard is a fraction of what it should be, giving falsely large values for analyte (the actual concentration is confirmed via IC). I should note this is not always the case. Depending on where the sample is from there may be no issues, I do the injection and the results are reasonable. But for certain others, if the results were accurate, I should be able to smell it across the room.
I've recently switched from crotonic acid to pivalic acid and the problem continues.
I know GC FID, water, and VFAs should never really go together, but this machine was purchased before I came to the company, and we'd all rather 'fix' these problems as best as possible instead of returning to paying someone else to do them on an IC.
Thanks for the insight,
Jessica
My parameters are:
Focus GC FID
Column; Stabilwax-DA column (Restek 30m-0.25mm ID-0.50μm df)
Four Temp; 100°C to 190°C @ 5°C/min, 240°C @10°C/min
Inj. temp; 200°C split 7:1
Carrier; N2 constant pressure 225kPa
Det. Temp; 200°C
The problems are these, ghost peaks on blank injections and decreased standard peak area for samples as compared to calibration injections.
Problem #1 a blank of water has peaks for all the analytes and the standard. running with machine without a physical injection shows no peaks. The ghost peaks seem to be a problem of residue somewhere, but where? If I replace the septum, liner, syringe, solvent vials and solvent, it goes away. But with just one injection (10-100ppm analyte) they're back. It is not economical for me to replace all this each injection. I did some calculations and my vapor volume is 1/3 that of the liner volume so I'm doubting backflash is causing the residue. I'm rinsing the syringe 5 times with water and 5 times with acetone/methanol before and after injection and three times with sample. The peaks will minimize if I inject water many times, but who has the time for that? Ideally I'd like the analytes and standard to stop sticking to whatever they're sticking to.
Problem #2 I can run several calibration standards and get a pretty decent calibration using the internal standard method. The peak area is pretty stable. Then I inject a sample and the standard is a fraction of what it should be, giving falsely large values for analyte (the actual concentration is confirmed via IC). I should note this is not always the case. Depending on where the sample is from there may be no issues, I do the injection and the results are reasonable. But for certain others, if the results were accurate, I should be able to smell it across the room.
I've recently switched from crotonic acid to pivalic acid and the problem continues.
I know GC FID, water, and VFAs should never really go together, but this machine was purchased before I came to the company, and we'd all rather 'fix' these problems as best as possible instead of returning to paying someone else to do them on an IC.
Thanks for the insight,
Jessica