Well lack some detail like the GC setting, column split flow etcetera, but will have a go at the headspace parameters in general.
The Headspace settings that I am using are:
Incubation Temp 85C- might be too high especially if water is present in significant amounts, vapor may spoil sensitivity and precision, usually try to avoid above 75-80°C with water
Transfer line 110C
Loop 110C
T-line and Loop, well hard to tell depending on whats in the sample and if any polar high boilers may be present, as a precaution as you have unknows I would consider raising the considerbly like 180/190°C respectively to avoid any contamination of valve/loop/needle
Incubation time 60min-fair enough as you are screening but can be reduced when you have better info
Agitation high-do not hurt
Vial pressurisation 0.4min-What is the setting for vial pressure?
Loop fill 0.08min-In my experiance set to short for Agilent instruments below 0,15min you may have issues 0,20min should do better
Loop equilibration 0.2min-depend on loop size if standard 1ml I would set it to 0,05min
Sample inject 0.5min-Normally OK but depends on how your total flow looks like versus loop size, if for instance it is a EPC set up where the total flows go through HS and splits up like column 2ml/min and split 10ml/min the loop is flushed OK.
Oven stabilization 1min-OK
As for RSD% in general I would expect a RSD below 10%, pharmacopiea genereal procedures tend to accept a RSD of 15% which might be OK for trace limit test analysis of like 2ppm benzene but at 500-1000ppm level of toluene which have a good response on FID I would expect better.
Using water on top on the sample. Well I have mostly seen failures regarding water and toluene performance wise, technically you can dissolve some toluene in water but considering your settings you might have quite unstable partion coeffient bettwen liquid/gas phase (more or less completly forced to gas phase and the vapor issue). You could consider using an aprotic solvent such as DMF/DMSO unless those are likely to be present, that way you can keep headspace oven at like 70-80°C and have a more stable partion coefficient as those sample solvents retains toluene better.