PEST DEGRADATION- EPA8081A
Discussions about GC and other "gas phase" separation techniques.
7 posts
Page 1 of 1
Please Help me to over come the Endrin/DDT degradation problem.give me some tips. i changed my column , inlet liner,gold seal,clean the injection port.please give me the conditions of GC. I have agilent 7890a gc with micro ECDs.
-
- Posts: 89
- Joined: Fri Aug 15, 2008 11:44 am
What are the conditions you are operating now? To what extent the compounds degrade?Please Help me to over come the Endrin/DDT degradation problem.give me some tips. i changed my column , inlet liner,gold seal,clean the injection port.please give me the conditions of GC. I have agilent 7890a gc with micro ECDs.
-
- Posts: 3
- Joined: Tue May 11, 2010 6:54 pm
At my laboratory our DDT/Endrin breakdown is usually < 3% but we have to change the inlet liner to achieve this. We use a liner with glass wool and that is usually the only maintenance required to meet the breakdown criteria. It's a little more expensive but relatively easy to spend about a minute changing a liner vs cooling down the instrument and clipping columns and replacing gold seals.
-
- Posts: 14
- Joined: Tue May 11, 2010 9:40 pm
Watch your injector temp. >197 will cause significant breakdown. Clean and silanize your liners and if you use glass wool, make sure it's not crappy glass wool. Make sur your injector temp is being read properly too, I've found variations in temperature probes. Just for fun, put a thermometer in the injection port and see what you get.. I've seen a few surprises. If these things don't work, shoot me an e-mail.
Good luck,
Bill
Good luck,
Bill
-
- Posts: 16
- Joined: Fri Feb 05, 2010 5:36 pm
At my laboratory our DDT/Endrin breakdown is usually < 3% but we have to change the inlet liner to achieve this. We use a liner with glass wool and that is usually the only maintenance required to meet the breakdown criteria. It's a little more expensive but relatively easy to spend about a minute changing a liner vs cooling down the instrument and clipping columns and replacing gold seals.
Can you please give me the part number for the liner.
-
- Posts: 16
- Joined: Fri Feb 05, 2010 5:36 pm
Watch your injector temp. >197 will cause significant breakdown. Clean and silanize your liners and if you use glass wool, make sure it's not crappy glass wool. Make sur your injector temp is being read properly too, I've found variations in temperature probes. Just for fun, put a thermometer in the injection port and see what you get.. I've seen a few surprises. If these things don't work, shoot me an e-mail.
Good luck,
Bill
Instrument 1: Agilent 7890A Gas Chromatograph
Detectors: Electron Capture Detector (ECD)
Auto sampler: Agilent 7683
Column 1: RTX-CL Pesticides 1, 30m x 0.32mm x 0.32µm
Column 2: RTX-CL Pesticides 2, 30m x 0.32mm x 0.25µm
Flow Rate: He @ 4 mL/min.
GC Conditions: Initial Temp: 120oC
Initial Time: 0.3min
Injector: 200C
Det:330C
Rate A: 45o/minute
Temp. A: 200oC
Time A: 0.0 min.
Rate B: 15o/minute
Temp. B: 230oC
Time B: 0 min.
Rate C: 30o/minute
Final Temp: 320oC
Final Time: 2.0 min.
-
- Posts: 16
- Joined: Fri Feb 05, 2010 5:36 pm
What are the conditions you are operating now? To what extent the compounds degrade?Please Help me to over come the Endrin/DDT degradation problem.give me some tips. i changed my column , inlet liner,gold seal,clean the injection port.please give me the conditions of GC. I have agilent 7890a gc with micro ECDs.
more than 30% degradation
nstrument 1: Agilent 7890A Gas Chromatograph
Detectors: Electron Capture Detector (ECD)
Auto sampler: Agilent 7683
Column 1: RTX-CL Pesticides 1, 30m x 0.32mm x 0.32µm
Column 2: RTX-CL Pesticides 2, 30m x 0.32mm x 0.25µm
Flow Rate: He @ 4 mL/min.
GC Conditions: Initial Temp: 120oC
Initial Time: 0.3min
Injector: 200C
Det:330C
Rate A: 45o/minute
Temp. A: 200oC
Time A: 0.0 min.
Rate B: 15o/minute
Temp. B: 230oC
Time B: 0 min.
Rate C: 30o/minute
Final Temp: 320oC
Final Time: 2.0 min.
7 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science