I am doing chlorinated pesticides assays on Agilent 6890-ECD. Pair of chromatographers. Detectors (mikro ECD) are quite old ones (~ 6years). Probably they have been dead by now, but we do not have many samples. Gas carrier is N2. Problem is just one annoing ghost peak on boths colums (HP5, DB35MS). And most annoing of all — those ghost peaks acts as one of my sreened pesticides. And reacently it began to act as one of PCB (diferent assay, but same owen and column method), that newer happened before. Interesting thing is I do not see those ghost peaks while chronographing standart solution or blank solution. It only happens during sample sreening.
Things tried: baking of colums, twice, heating of ECDs (many times for another reasons), colums trimming (twice already), inlet cleaning (really — washing with solvent like petroleum ether, n-hexane, acetone). Have tried working surfaces cleanliness, working solvents, blank cleaning gels, cleanliness of equipment that is used. Didn't find the source.
Last thing. Our equipment have been used with splitless liners instead of spilt ones for ~5 years (in split mode). So could happen kind of bild up of fatty residues somewhere, but colums were trimmed and inlets washed, liners, septas, ferules, goldplated seals changed.
Any ideas?
Thanks.