Calibration curve linearity issues and loss of sensitivity
Posted: Wed Apr 28, 2010 6:48 pm
Good afternoon,
First off great job to all on this site... you guys have helped me alot.
Here's my prob...
What I'm running... 6890 gc with 5973 msd ei, column you're standard db-5, 30m, 0.25mm, 0.25um
performed a full maintenance on the system... changed liner (single tappered with glass wool, new septum, gold seal, cleaned the source, changed the filament.
we also added a guard column.
Leak test good, performed autotune... EM volt went from 2250 to 1775 and my abundance for all ions are whithin range.
I've prepped a 5 pt calibration ranging from 30 ppm to 200ppm... outcome used to be very linear but now it's quadratic.
I've cut a section of the guard colum... no change, I've redid my stds (3 times)... no change.
Here's what I'm thinking... column was installed last summer so most likely it has active sites... if this is the case shouldn't I see some tailing, which i don't.
There was mention on this board about changing the EM... the EM on this machine has never been changed
Could the guard column be a factor?
As for sensitivy... if my ions during autotune look ok but the std output seems low what could this be?
Thanks in advance
Syl
Here's my prob...
Thanks in advance
Syl