We use a Varian-CP 3800 (Column type Porapak QS, 80-100 MESH, Dimensions: 4mx1/8"x2.0mm SS; split valve at the end of the column that separates flow between ECD and FID). Nitrogen is used as carrier gas (nitrogen generator). FID settings: hydrogen 30ml/min, nitrogen 30ml/min, air 300ml/min, detector 300ºC, oven 30ºC, injector 70ºC.
Problem: The FID is showing an additional peak exactly overlapping the CH4 peak (in standards and air samples). This additional peak is separated into a strong negative and a less strong positive part; furthermore the peak is drifting out at the end so that the end of the peak area is hard to define. The negative part of the additional peak is getting less strong over time but is not disappearing. Repeated analysis of standard gases shows high standard deviations (standard deviation for 5 samples between 0.21 and 1.37).
Changes that were made so far: Change in parameters like injection volume, injection speed, column temperature, injection temperature, air flow and flow rate. The column and detectors were baked for cleaning. The flame tip was checked. None of these changes did help.
Any ideas?
Many thanks!