Hi,members
There are several questions about RT puzzled me:
1,Many factors could effect RT,for example, RT in AM may be different from PM, so what is the accept criteria of the RT variation?
2,Is there any difference of the RT accept criteria between assay and impurity?
3,Where is the accept criteria from?
Thanks a lot, any comments would be welcome.

As with so many things, "it depends". The retention time must be reproducible enough to avoid misidentifying a peak. If it's a simple assay, (only one peak, for example), the retention time can vary by several percent and not cause a problem. If it's a complex sample, then retention times would need to be closer.

That said, look at the specification for flow rate accuracy/reproducibility on your pump. That will put a lower limit on the retention time reproducibility for your chromatography.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 54 users online :: 0 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 54 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry