Recently I have tried to set an old HPLC method for the determination of Ibuprofen in a new system. My surprise comes when I check the instrumental detection limit. It is about 10 times higher than before!
I have played with DAD settings of the new system to improve the detection limit, with little success.
In my old lab I work with an Agilent 1100 series system with a UV-DAD 1100 detector. The column was a Zorbax eclipse DB-C8 4,6 x 150 mm column, particle size 5 um.
Now I work with a new Jasco system with MD-2010 Diode array detector. The column is a Kinetex C18 4,6 x 100 mm , particle size 2,6 um.
The method I use is the same is the same except the flow rate. For Zorbax column the flow is 1 mL/min while for Kinetex column 2 mL/min. Isocratic conditions 50:50 ACN:Buffer pH=3. Injection volume 20 uL, column temperature 40 degrees.
While in the old method (Agilent system with Zorbax column) I get an instrumental detection limit (IDL) of 22 ng/mL with the new system I obtain an IDL of 320 ng/mL. IDL was calculated as the concentration giving 3 times the baseline noise.
My first reaction to this difference was to prepare the standard solutions again from the beginning, but I get the same results. I should also mention that other compounds in the method show good IDL between 10-20 ng/mL, except for clofibric acid which has an IDL of 60 ng/mL (another method with agilent system game me an IDL of 35 ng/mL)
So my question is, if an instrument brand could give such big differences in IDL. If not…any idea why I get this big difference?
Thanks!