The maximum pressure that supports the new column is 400 bar and was fitted with the mobile phase phosphate increased gradually flow of 0.5 ml / min up to 2 ml / min and the pressure varies from 116 bar to 365 bar. For a second use the pressure reached the limit and shut down the HPLC.What was the maximum pressure you saw when the column was new?
And did the increase in max pressure occur gradually or abruptly?
The methodology used is of Agilent recommended for Zorbax Eclipse AAA column. We have done the test off-line in a graduated cylinder and there is not precipitation when mixed mobile phases.Do you use on-line mixing of the solvents?
From your posted mobile phase composition, is it possible that 40mM phosphate buffer is precipitating in the organic rich mobile phase after mixing (and thus causing column blockage)? Have you tried to mix all the solvents off-line in a graduated cylinder?
Just my 2$
We will conduct further tests at lower concentrations of buffer and changing the flow to reduce pressure. The methodology used is found at the following addressWhen doing precipitation tests of mobile phase one should be aware that one can get supersaturation. A supersaturated solution can precipitate when hitting a nucleating system like a column frit.
I prepared a phosphate-MeOH buffer once which was completely clear and it gave a surprisingly good performance. After a few weeks I noticed nice needle crystals in it. These very readily dissolved in water, I was sure that some phosphate had precipitated. I never used it again, not wanting to base the chromatography on luck. Incidentally, the more basic the phosphate the higher the tendency to precipitate, your 90% organic with the 7.8 pH phosphate looks very suspect to me, even at that concentration. I don´t remember what the composition was of my mobile phase, but it was certainly less organic, though most likely at 0.1M.
Thanks, we will conduct further tests by change of the composition organic mobile phase, taking into account the link sent.Here is an article on buffer solubility: http://chromatographyonline.findanalyti ... rticle.pdf