Impact of UV detector restart on the Response

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hi, all,

I was run the same thing (everything is same in term of concentration, injection volume, instrument method, etc) before and after shutdown and restart the instrument together with UV detector. The response (in term of area) showed 6 times difference. is this possible or any other reason explain this? thanks a lot.
emily lee

Was the change in area a decrease or increase? Were there any other changes that you noticed? Per your post, the 'shutdown and restart' seem, at least to you, to be involved in this change. Was there anything that precipitated this event? Did something cause you to need to restart the instrument? Did you notice anything odd before or after the restart? Have you re-prepped and re-run standards and compared them to previous injections to determine if this change in area is a one-time event?

If you had, for instance, a power outage, and you had not saved some important changes to a method (specifically, injection volume), then upon restart the instrument may have loaded the old version of the same method you had been running, but with the previous injection volume (say, 100 uL vs. 20 uL). Just a thought. You may want to check each parameter to make sure it really is "the same as" before the shutdown.
Time flies like an arrow. Fruit flies like a banana.

Thanks for reply.

Nothing changed in the method and sample except the system restarting. I compared the instrument method attached to each specific injection.

I ran the same sample with all the setting the same, it looks good. Then I decided to repeat with more injections to confirm the method. I restart the system due to the system looks not fully communicating with the computer. I expect area change after restart the system, but not so much as 6 times more.

It is even better, but just curious about the reason.
emily lee

How old is the detector lamp and how long after restarting did you inject the samples?
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

meillid, why did you expect an area change after restarting the instrument?
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry