GC performance
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
Hi there, I did inject a 100 ppm of methyl estrate with scan mode I got an area of 409003309 and with SIM mode I got 132228767. Can some one tell me if my GC MS has a good sensitivity. Faris
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
The sensitivity specifiction for GC/MS will be a signal to noise measurement of a few picograms on column, with a specific column, flow rate, temperature profile and acquisition rate. Also make a blank solvent inejction between injections of high concentration solutions of ethyl sterate and low concentration solution. Sterates tend to carry over between injections.
Eyeball estimate, you have a shot at a visible peak at 10 ppb and even 1 ppb. So as long as the noise is low, you may be in the ball park.
Eyeball estimate, you have a shot at a visible peak at 10 ppb and even 1 ppb. So as long as the noise is low, you may be in the ball park.
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science