Dear friends...

I bought a GCMS QP2010 Shimadzu about one year ago and the turbomolecular pump works but the higher vacuum is not reached. I read another topic in this forum that presents another friend with the same problem, i tried all the suggestions on that topic but the problems persists. Anyone knows what is the real problem. I have seen others posting on forums the same problem but i haven't see none progress. Anyone knows what is this problem?

Thanks...

What are your column dimensions and flow rate, and what vacuum are you achieving?

2.0e+001 Pa (pressure achieved until shutdown)

1 ml/min to 1.8 ml/min (flow rate that vacuum starts). Mostly 1.8 ml/min

Column dimensions: 30m x 0.25mm x 0.25 um

That looks like a Pirani gauge (low vacuum) reading, not an ion gauge (high vacuum) reading. If the Pirani gauge reading does not go below 2.0e+001 you have a large air leak into the system. If the system is leak free the reading should go down to about 5e+000. The most common places for a large leak are the column interface to the mass spectrometer and the front opening door on the mass spectrometer. Try re-installing the column with a new nut and ferrule, and when you pump down the mass spectrometer I would recommend not tightening the screw on the mass spectrometer door and push the door against the O-ring after the rough pump starts and let the vacuum draw the door against the O-ring evenly. I have seen leaks when the screw is tightened before pumping down and the door is not sealing evenly around the O-ring.
5 posts Page 1 of 1

Who is online

In total there are 45 users online :: 1 registered, 0 hidden and 44 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 44 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry