Hello, I am another new bird on GC in this forum. Now I am trying to develop a GC method to seperate two diastereomers of alcohol. MW is about 100. They prefer to use regular GC column instead of chiral GC column. I think it is possible to use regular column because diastereomers are quite different in properties not like enantiomers. But I completely no idea how to select GC column. Anybody has any suggestions where to start with? Thank you so much.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Low MW Underivitized alcohols elute with symmetry and efficiency on only a few phases, the best are probably PEG type phases.
PDMS phase columns may work for you. Probably the better solution in any case is to have as much separating efficiency as possible, IE thin film and as small a diameter of capillary column as possible.
Good luck,
Rodney George
consultant
PDMS phase columns may work for you. Probably the better solution in any case is to have as much separating efficiency as possible, IE thin film and as small a diameter of capillary column as possible.
Good luck,
Rodney George
consultant
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
mibobdog,
You might take a look at the blood alcohol columns from Restek and J&W.
Best regards.
You might take a look at the blood alcohol columns from Restek and J&W.
Best regards.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We use either PEG and 624 type capillaries for lower alcohols. 624 is used in USP 611.
-
- Posts: 621
- Joined: Tue Nov 27, 2007 1:20 pm
You'll find many of the major supplier have selection guides.
Phenomenex also do a column selection by manufacturer in their catalog.
Phenomenex also do a column selection by manufacturer in their catalog.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
-
- Posts: 200
- Joined: Tue Apr 20, 2010 12:31 pm
We use a ZB-624 for separating methanol, ethanol and IPA almost everyday.
30m x 0.53, DF 3 i think. Resolution of 4 usually.
30m x 0.53, DF 3 i think. Resolution of 4 usually.
-
- Posts: 2
- Joined: Mon Apr 26, 2010 6:13 am
thank you all for all of your suggestions. except peak tailing, I also concern about if diastereomers can be seperated in regular GC column. I will try and post the update here.
thanks again
thanks again
7 posts
Page 1 of 1
Who is online
In total there are 89 users online :: 0 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 89 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 89 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science