Hi people!

I work at GC 6890, I analize pesticides in fruits and vegetables.

The standard has 30 pesticidas, and that includes HCH (α, β, γ, δ).

I noticed gladually a shift of HCH β and γ: now, those co-elute, but the retention time of the others pesticides are ok.

What can the problem be?

Thanks

First guess is that you have run many injections on the column - and the residual junk from injections has changed the stationary phase.

(I am also assuming that you are keeping up with changing liners, septa and such.)

Assuming there have been no other changes in the column and column conditions, like column length, gas flow rate or temperature program, try baking out the column. If that does not work, trimming off the front end of the column may work. And, if either of these techniques works, be sure you have a new column on hand or on order. The column's days are numbered.
2 posts Page 1 of 1

Who is online

In total there are 89 users online :: 0 registered, 0 hidden and 89 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 89 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry