Agilent 1100 drifting retention times
Posted: Wed Apr 14, 2010 6:31 pm
Is there any experience with retention times drifting when using an Agilent 1100 binary pump with solvent switching valve? We are running a method that we have been using without change for 10 years. Mobile phase is PO4-TEA buffer/ ACN, column is Zorbax C18. The gradient runs from 10% ACN to 70% ACN over 17 minutes, post-column (re-equilibration) time 5 minutes. We have an Agilent vacuum degasser prior to the pump.
We have recently had problems that retention times will increase on subsequent injections (to about 4 minutes late). I have changed seals, piston, outlet ball valve. Measured flow rate is correct, and the flow appears to be consistent through the inlet tube, judged by pulling up a small air bubble into the line and observing that it does not pull backwards in the line with pump stroke. Changing B side to 90% ACN/10% water seemed to correct the problem, but changing parts did not correct.
Thanks in advance!
We have recently had problems that retention times will increase on subsequent injections (to about 4 minutes late). I have changed seals, piston, outlet ball valve. Measured flow rate is correct, and the flow appears to be consistent through the inlet tube, judged by pulling up a small air bubble into the line and observing that it does not pull backwards in the line with pump stroke. Changing B side to 90% ACN/10% water seemed to correct the problem, but changing parts did not correct.
Thanks in advance!