Problems with Source heater assembly in MSD 5973N

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Two questions for discussion

Has anyone ever had failures with the ion source heater sensor?

Do you consider it a changeable piece in your equipment? If so, does anyone have a schedule for it?

ALC,

In all of my years working on people's 5973's, I have never changed a source heater or RTD. Therefore, I would not put it on a schedule. I would not even keep a spare around based on my experiences. Hope that helps...

Best regards.

I keep several RTDs and heater cartridges on stand by. I change 20 ion sources a month and eventually the wiring will break.
Try Scientific Instrument Services which sells both. Talk to Tony if need be... A screw driver or small allen wrench is all you need.

I do keep a spare on hand. But it happens so rarely it doesn't seem to need scheduled replacement. With 14 GCMS's, I've only had to replace maybe a total of 5 in the past 8 years. They're not the cheapest items either. Just keep a spare around and be careful not to bend the wires much during normal maintenance.
4 posts Page 1 of 1

Who is online

In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Bing [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry