GC/MS column contamination
Posted: Mon Apr 12, 2010 6:36 pm
by faris
Hi everyone. I am looking for a procedure to bake my column. I have a HP-5
in a 5970/5890. what would be the injection, detector, and the oven baking temperature and for how many hours.
Posted: Mon Apr 12, 2010 11:29 pm
by Don_Hilton
If you have just installed the column, there should be instructions with the column for conditioning it (and some of us don't bother with that these days, as columns come pretty clean.)
If you are trying to bake off residue from an analysis - it depends on what it takes to get rid of the residue. And, if you are trying to get rid of column bleed, baking may just make things worse.
Inlet temperature: There are two schools on operation of the GC. Some of us run the inlet a bit cooler than the maximum temperature used in the analysis - assuming that it is easier to replace an inlet liner than a column and it takes more temperature to move junk across a stationary phase than across a liner (even with glass wool). Others run the inlet up to be sure everything gets onthe column. - So the first question is how have you been operating the instrument? If you have beek keeping the inlet cool, change the inlet liner and keep the inlet cool. If you have been running the inlet hot, change the liner and keep the inlet hot. (Clean the inlet by changing the liner - not by pushing garbage through the column)
The column - If you have been running an analyisis with an upper temperature on the column that is below the temperature required to elute some copunds, run the temperature high enough to elute the compunds - and do it every run rather than collecting stuff on the column. That stuff becomes part of the stationary phase for the next injectoon(s) And, as the stationary phase changes, so does your separation.
If you have materials that will not elute in reasonable time at the upper limit of your column. You may need to remove a portion from the front end of the column. For some, this works. For the work I've done, I've discovered that it is more effective to simply replace the entire column. (When I kill a column, I kill it dead!)
Detector temperature - this scares me a bit. If you will be pushign through heavy stuff that leaves you needing to raise the temperature of a mass spec to avoid fowling -- expect ot have to clean the ion source anyhow. And this gets into the question of how long to bake: Not long. And, you will need to moitor what is coming off to know when to stop (thus detector temperature close to normal operating temperature - or just a bit warmer).
Run the column up in temperature to close to the max. If you think there is a lot of stuff to bake off, hold it a while ( 10 min to 1 hour - you learn this from experience with your analysis) and then make a run on your instrument - like an analytical run, but no injection. Watch the background. If it low enough, you baked the column long enough. If the background seems to be going up or will not fall, you may have a column that has reached the end of its life. If running the column up to close to the maxiumum and holding for 5 to 10 minutes at the end of a run is not enough to keep your column in shape, look at your sample prep -- you are putting things on your column that will kill the column.
Posted: Tue Apr 13, 2010 1:04 am
by faris
Thank you Don, What I am thinking to do is elevate the column temperature to 300C. the injector to 300 C and the detector to 300 C meanwhile I will disconnect the column from the detector interface and increase the flow rate to 30 mL/min. do you think this will help???
Posted: Tue Apr 13, 2010 2:58 am
by haiedc
Thank you Don, What I am thinking to do is elevate the column temperature to 300C. the injector to 300 C and the detector to 300 C meanwhile I will disconnect the column from the detector interface and increase the flow rate to 30 mL/min. do you think this will help???
How come you set the column flow to 30ml/min? It's not necessary that high anyway.
Posted: Tue Apr 13, 2010 5:51 am
by faris
I usually increase the Flow rate to 30mL/min in order to remove all the contaminants, I think the high flow rate in necessary for removing them.
Posted: Tue Apr 13, 2010 11:15 am
by Don_Hilton
If you remove the column from the detector, you have no way of monitoring to see if the column is coming clean.
The column temperature you list is OK. The flow rate is quite high.
Clean the detector by the normal method for cleaning a detector. For an FID, there is a set of brushes available. Pull the detector apart and clean with the brushes. If you run the detector at 300 degrees, then 300 degrees can be OK - just be sure that it has gas flow so that it is not fouled by prolonged heating without colulmn flow.
Beyond that, the conditions you use are entirely dependant on your samples and method - of which we know nothing.
Posted: Tue Apr 13, 2010 12:58 pm
by Ron
Do you mean the column flow rate or the total through the inlet? If you are using a standard 30 m 0.25 mm id column you would have to have a head pressure in excess of 100 psi to reach 30 mL/min. I usually use 2 to 3 mL/min column flow during a bakeout of a 0.25 mm id column.
Posted: Tue Apr 13, 2010 1:34 pm
by faris
Hi Ron. What would be the head pressure for 1ml/min and also what it would be for 2ml/min. About disconnecting the column for the detector, After the baking process is finished and reconnect the column and run a blank injection to monitor the base line and if the contaminations has been removed.
Posted: Wed Apr 14, 2010 8:29 am
by Peter Apps
Why do you want to bake the column - baseline drift with temperature programming ? There are two main causes for basline drift as columns heat up; accumulated heavy contaminants, or column bleed. If you have heavy contaminants you need to reverse the column during baking so that the detector end is in the inlet, otherwise all you do is to push heavy muck further into the column without eluting all of it. If column bleed is the problem baking will only make it worse.
In either case you need to unsure that your carrier gas is free of oxygen and water vapour by installing scrubbers in the gas line before you take the column anywhere near its maximum temperature.
Peter
Posted: Wed Apr 14, 2010 1:20 pm
by faris
Hi Peter. I do reverse the column while baking but I don't connect it to the detector. I just put a septum on the detector inlet to keep the vacuum stable.