I have a problem,
I'm making some repeatability test on my GC system.
I'm making an headspace analysis using an autosampler.
The speed of the gas is 35cm/sec; the split is 1:250. The temperature of the column (unpolar, 30m) is 170°C.The detector temperature is 300°C and the injector temperature is 250°C.
I'm injecting 3methil penthane (iso octane). I'm using microcaps for the sample addition in the vial.
My problem is that as concern peak height I have a CV% of 0,2-0,8% (that for me is ok), while the area CV% is between 1 and 2,5% (that for me is not ok).
Can you give me an explanation of this phenomenum?
I want to have a CV<1%.
Thanks
Greta