Signal to noise ratio

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Is there any way to automatically calculate signal to noise ratio with Agilent ChemStation (HPLC) ? Thanks.

Yes there is. Load your data into the Data Analysis view, from the 'Report' drop down menu select 'System Suitability' and then 'Edit Noise Ranges'. Enter your noise range in the boxes.
Print the built-in 'Performance + Noise' report to see the result.

Be careful which numbers you report though. From memory ChemStation does h/H rather than 2h/H (may be the other way round). I will double check when I get a chance to go through the manual.

Hope this helps.

Mike
Insanity: doing the same thing over and over again and expecting different results.
Albert Einstein, (attributed)
US (German-born) physicist (1879 - 1955)

Also, I believe the calculations are only done for named peaks, so make sure you at least have the retention time in your calibration table.
Merlin K. L. Bicking, Ph.D.
ACCTA, Inc.

You are both correct; I tried it and it worked. And I made my own cheat sheet for the future for when I forget how. Thanks.
4 posts Page 1 of 1

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry