Detection of oxygen with TCD
Posted: Wed Apr 07, 2010 12:04 pm
WE are having one NGA having the configuration of Porapak and molecular seive columns and TCD. We are facing a unique situation in the detection of oxygen. If we inject (manually) room air the peak area ratio for the first injection comes as oxygen (15%) and nitrogen (85%). for next air injection the ratio changes to 19% and 81% respectively. for the subsequent injection the ratio remains constant. However if we inject a natural gas (C1 to C8) analyse it and then again inject room air the peak area ratio reverts back to 15 % and 85 % .
Is this problem related to columns or TCD?
What is the maximum conc of oxygen that can be detected by TCD accurately?
Is there any solution to this problem?
Is this problem related to columns or TCD?
What is the maximum conc of oxygen that can be detected by TCD accurately?
Is there any solution to this problem?