sample sets do not work. Who can help me?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
Hello, I am using the Empower software. I build a sample sets, however, the injection could move automatically from vial to vial. For example, my run time is 10 minutes, after ten minutes I noticed that at the bottom of the screen showing "Injection running" and after about 10 minutes showing "intrument failure". So everytime, I have to mannually choose "Abort and move to the next line" to continue my injections. Does anybody meet this problem before?
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Smartcici,
Most probably, the trouble is caused by an invalid instruction in the instrument method.
Open the instr. method and find out what happens at “time 10 minâ€
Most probably, the trouble is caused by an invalid instruction in the instrument method.
Open the instr. method and find out what happens at “time 10 minâ€
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 13
- Joined: Mon Mar 06, 2006 8:19 am
Hi all,
I also have got a problem like smartcici.
I using Empower2 with UPLC and I quite sure my instrument method don't have anything wrong. I cann't inject sample in sequent both gradient and isocretic method.
Please help!
I also have got a problem like smartcici.
I using Empower2 with UPLC and I quite sure my instrument method don't have anything wrong. I cann't inject sample in sequent both gradient and isocretic method.
Please help!
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Kasiban,
More detail is needed. Especially the exact events, as you experience them.
F. x. what error messages do you get – if any? Does the system stop after exactly after10 min?
So, help doesn’t always come by its self – sometimes it takes an effort.
Best Regards
More detail is needed. Especially the exact events, as you experience them.
F. x. what error messages do you get – if any? Does the system stop after exactly after10 min?
So, help doesn’t always come by its self – sometimes it takes an effort.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
Did you tried to make an new instrument method?
Not "save as", but make a new instrument method, based on your system.
We had this problem in the past and after making a new instrument method, the problem was gone.
Ace
Not "save as", but make a new instrument method, based on your system.
We had this problem in the past and after making a new instrument method, the problem was gone.
Ace
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Whenever confronted with something like this, it is a good idea to also post whatever is in the Empower Message Center as well.
Thanks,
DR

DR

6 posts
Page 1 of 1
Who is online
In total there are 125 users online :: 3 registered, 0 hidden and 122 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 122 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 122 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science