A couple of things:
1) my comments about getting things to chromatograph was based on the term "neat oil." My mind wandered off to neatsfoot oil or oil from the neat tree... (This is what I get for logging on the computer before breakfast)
2) With a flow of 0.5 mL/min going through your column, you should be able to see peaks from your mixture, if they are there. While optimization should help, this is not the big sensitivity problem
As the others have suggested a faster ramp rate may help out. I would try more material on column and faster chromatograpy and then back off to separate the components that are of particular interest.
The next question is what is your injection volume? While you have some film thickness to help - it is more than easy to overload a 0.1 mm ID column. And, the volume of solvent is part of the loading. Is it a problem? I would suggest ramp rate, quantity of analytes, and detector optimization first.
I would suggest that you dilute the crude in carbon disulfide (which has no signal in the FID) at enough dilution that you do not have a viscosity problem in the syringe and then use a 5 microliter syringe - or smaller and inject as small a volume as you can with your autosampler. Perhaps 0.1 uL.
And use a very small volume liner. At the low flow rates, it takes a long time to get everything on the column (and too short a purge off time will really hurt!). A smaller liner with a smaller solvent volume - to avoid overfilling the liner with vapor takes less time to transfer to the column - and you avoid band broadening on the front end.
Using a 0.2 mm ID colum, I've injected small volumes (0.2 or so uL) of naphthas (neat) to profile the mixtures. I wanted solvent free injections because I was using a mass spec...