I'm working with Chemstation methods that were created by someone else.....and the reports read Norm% values. Is it possible to customize the report to read wt%, mol% or vol%?
Our Standard method and Sample method are identical (except the response factor from the standard method has been applied to the sample method.)
Our standard certificate results are measured in volume%. So it was those certified values that were entered into the Standard method, and the response factors from the standard run were used in the creation of the sample method. The certified volume% values match the report Norm% values for our standard runs. Now is it safe to say that the Norm% values obtained from the sample runs would then too also be vol%?
I tried running a different standard which certified results were given in mol%, and the Norm% report values did not come close to matching.
The standard is a 8 component Liquid mix piston cylinder (methane to pentane & C6+ Balance). The samples are unknown volatile liquid mixtures of these components.
Any suggestions?