Working on a headspace method here for the analysis of ethanol in a pharmaceutical suspension. Samples are prepared in de-ionized water, with isobutanol as an internal standard. GC is a Shimadzu 2014 GC with an CTC autosampler (syringe) and FID detection.
The problem we're encountering is %RSD failing for linearity and accuracy preparations. Criteria is set at 10% RSD, which is pretty large, yet %RSD values are all over the place.
Linearity is from 2.5 ug/mL to 1000 ug/mL. Vials for each level are prepared individually, with internal std added (125 ug/mL isobutanol). For linearity, we do 3 injections of each level.
Accuracy samples are prepped a bit differently. Suspension is measured out (1000 mg) into a 10-mL volumetric flask, and spiked with ethanol to concentrations of 250, 500 or 750 ug/mL. From these spiked solutions multiple vials are prepared with internal std and injected once each.
GC conditions as follows:
Column: Zebron ZB-624 column, 0.32 mm x 30 m, 1.8 um layer of phase
Carrier gas: He @ 35 cm/sec
Split ratio: 1:10
Injector Port Temp: 110 C
Detector Temp: 250 C
Column Temp Program: 40 C for 1 min, ramp to 80 C @ 15C/min, ramp to 240 C @ 50C/min, hold 240 C for 1 min
Injection volume: 1.0 mL headspace
Headspace conditions:
Incubation Temp: 80 C
Incubation Time: 7.5 min
Syringe Temp: 105 C
Agitator Speed: 300 rpm
Any suggestions are most welcome. I've already tried changing liners, septa and even replacing the injecting syringe.