You only really have the choice of gaseous or compressed liquid sample if you have methane present.
I certainly would be using pressuried valve sampling if you are trying to quantify them at % concentrations, and probably even at ppm concentrations.
The concern with any sampling that doesn't load a complete and representative sample onto the column is that there is a selection process occurring in the injector or sampling system.
In the case of split/splitless the more volatile components ( eg methane or permanent gases ) may be favoured over the less volatile ( eg pentane ).
It's possible to use split/splitless injectors, but you have to watch for discrimination, usually by analysing standards of differing composition at the extremes of your product range.
Any calibration is dependent on the split system behaviour remaining constant for all samples and standards - a big ask, especially if sample volatility is different to standards.
The piston is designed to ensure there is no headspace, because headspace of such a pressurised liquid will have a different composition to the sample liquid.
For pressurised liquid sampling, high pressure, low volume, liquid sampling valves are commonly used to ensure the sample pressure and composition is maintained into the injector.
Bruce Hamilton