high resolution mass determination

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
should I isolate compound or just use lc column. which one will give me more accurate mass. thanks.

I would expect the accuracy of the mass to depend on the instrument used, the mass calibration, the stability of a reference mass signal and the signal strength for your unknown. What are the differences between running the compound on the LC and isolating the compound?

I know that my question probably is silly but I just want to know what is the "general rule" to do HRMS. but it seems there is no such a general rule.
as i understand, pure analyte with pump syring could well control signal strength and less interference, which will guarantee more mass accuracy. but it is also very time consuming.
I would expect the accuracy of the mass to depend on the instrument used, the mass calibration, the stability of a reference mass signal and the signal strength for your unknown. What are the differences between running the compound on the LC and isolating the compound?
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry