Page 1 of 1

Baseline correction in clarity

Posted: Mon Mar 29, 2010 1:03 pm
by sana_sankar
Today I downloaded the demo version of clarity and ran few demos available in the package.I was surpurised to see, it does not implement automatic baseline drift removal.I am wondering, why clarity has not implemented automaBaseline drift removal, which is a major feature in Data Processing.

Is it because, for a Gas Chromotographer, baseline drift removal is more accurate only with manual interference than using the automatic methods.

Posted: Wed Apr 07, 2010 2:08 pm
by ivanvins
Please could you explain in some detail, what the mentioned drift removal function should do?

To compensate for detector drift or baseline Clarity enables you to automatically subtract defined baseline chromatogram (Method Setup, Advanced tab), you could change this later on the Measurement tab in Chromatogram window.

In case you are interested in Clarity, you could join our discussion forum

http://forum.dataapex.com/

and put your suggestions in the Wishlist section.

Ivan Vins
DataApex